crwdns2911028:0crwdne2911028:0
crwdns2862678:0crwdne2862678:0
crwdns2895503:0crwdne2895503:0
Failed Circuit Check
There is a failure in the circuit check that is appearing.
Inspect for leaks
Inspect and tighten, if applicable, the Patient Circuit, Exhalation Valve Adapter, Emergency Relief Diaphragm, Pneumatic Tubing Inside the Ventilator, TSI Air/Oxygen Flow Sensors Assembly, Inhalation Flow Outlet, and the Oxygen Sensor for leaks and damage.
Dissemble
 Dissemble the exhalation valve to inspect the valve’s diaphragm and seal.
Clean or Replace
Clean or Replace the Emergency Intake Diaphragm and Inspect the Rubber Seal between the Inhalation Outlet Block and Mixer Block.
No Flow or, Low Flow from the Inspiratory Port During the Inspiratory Phase of a Mandatory Breath
The inspiratory port is receiving no flow or low flow during the inspiratory phase.
Inlet Regulator Failure
If you are experiencing the symptoms of no-flow or low flow during the mandatory breath, you should calibrate the air and oxygen regulators.
Servo Valve Failure
To check if the servo valve is working, you must also calibrate the air and oxygen regulators. If the calibration does not fix the issue of flow, you will need to replace the air and oxygen regulators.
Monitored Values for Exhaled Flows or Volumes Differ from Settings
The values from the flow and oxygen sensor appear different on the monitor than from the settings they were set to.
Occluded or Missing Bacteria Filter
Check to see if there is a bacteria filter.  The bacteria filter can be obstructed from its use if the filter is dirty. If there is not a bacteria filter or the filter appears to be dirty the filter should be removed and replaced with a new filter.
Dirty Exhalation flow sensor.
 The exhalation flow sensor is dirty and needs to be removed and cleaned. To remove and replace the exhalation flow sensor the following guide will show the steps, Exhalation Flow Sensor Repair.
Exhalation heater failure.
Locate the Bacteria filter. Then touch the filter with your hand. If it is not warm to the touch, then replace the Exhalation Heater.
Vent settings for the “Circuit Type” humidity setup are incorrect for the humidifier system in use.
Correct the “Circuit Type” selection in the Patient Set-up menu.
Ventilator Cannot Cycle, an Audible Alarm is on, and an Error Message is Displayed in the Message Window
There is an audible alarm with a specific error message displayed in the message window.
Uncalibrated Servo Valves
If an error message is displayed in the message window and the ventilator cannot cycle, calibrate the servo valves.
If this is not successful in fixing the issue, you will need to replace the servo valves. Refer to table 3-2 to translate the device alert messages. This table can be found at the end of this page.
Loss of Battery Power/ Internal Battery Not Charging
Poor performance of the battery such as loss of battery power and the internal battery not charging.
Battery Not Charged
If your battery is low is showing loss of battery the battery should be recharged.
Defective Battery/ Defective DC-DC Power Supply
If recharging was not fixing the solution the battery or power supply is defective.  To fix this the battery/ power supply needs to be removed and replaced.
Rylan Loader
crwdns2935283:004/09/20crwdne2935283:0
243 crwdns2915208:0crwdne2915208:0
Cal Poly, Team S7-G21, Paton Spring 2020 crwdns2935289:0Cal Poly, Team S7-G21, Paton Spring 2020crwdne2935289:0
CPSU-PATON-S20S7G21
crwdns2935297:036crwdne2935297:0
crwdns2944067:04crwdne2944067:0
I worked on an e360 ventilator at a government hospital over the weekend.
When the unit is powered on, it doesn’t complete the booting process, instead It
displays “CMOS checksum error”.
with an option to “Press F1 to continue, DEL to enter SETUP”.
anybody with an idea of how to provide a permanent solution? Pressing F1 completes the booting process and ventilation is possible.
But i will rather not have the message in the first place.
- oluwafemi.esho@gmail.com
Femi Esho - crwdns2934203:0crwdne2934203:0 crwdns2950251:0crwdne2950251:0
You must change the coin battery "CMOS" battery
Soporte Técnico Digimedical - crwdns2934203:0crwdne2934203:0 crwdns2950251:0crwdne2950251:0
I have a problem with the e360T ventilator, the machine has a Pressure sensor error. I replaced the analog board with a new one. The machine worked for 20 minutes and then appeared again Pressure sensor error. Please help me resolve this error Thank you very much
My email address: ducthanh@meditronicvn.com
Thank you very much
THANH
ducthanh - crwdns2934203:0crwdne2934203:0 crwdns2950251:0crwdne2950251:0
i have a problem with the e360T ventilato, i was change the flow sensor, even the flow sensor board, the jumpers, the CPU, i was checking the neumátic system for to confirm if was a fuga, i was try to change the valvule, all the cable and the problem is that always we have a error all time after that i made all mentionated y The machine show the same message, flow sensor error please i need your help.
Stifen J. Vargas - crwdns2934203:0crwdne2934203:0 crwdns2950251:0crwdne2950251:0
crwdns2936753:030crwdne2936753:0 50
crwdns2959996:0crwdne2959996:0 1,584
