Philips Emerson CoughAssist CA Series Testing Procedure
  1. Philips Emerson CoughAssist CA Series Testing Procedure, Philips Emerson CoughAssist CA Series Testing Procedure: crwdns2935265:01crwdnd2935265:01crwdnd2935265:01crwdne2935265:0
    • For CM models, the pressure knob is located on the left in place of the timing knobs pictured here.
    • Procedure can be found in the attached document.
    • Block the connector on the front of the machine using a rubber plug or similar item.
    • Check the pressure gauge. If it is not zero, see “pressure gauge reading is wrong” on the troubleshooting guide.
    • Set the inhale pressure knob fully clockwise, and the inhale flow and pressure knobs fully counterclockwise. On the CA models, set the manual/auto switch to manual, and set each timing knob to 1 second.
    • Turn on the power switch, checking that the light under it turns on. If not, see “indicator light won’t turn on” in the troubleshooting guide.
    • Turn the pressure knob slowly clockwise, noting that the speed of the motor increases steadily. Once it’s fully clockwise, allow the unit to warm up for 1 to 2 minutes.
    • Reset the pressure knob counterclockwise, then set the control lever to inhale and check that the gauge reads between 2 and 5 cmH2O.
    • Move the control lever to exhale and note that the gauge reads between -2 and -5 cmH2O.
    • Repeat the inhale and exhale tests with the pressure knob fully clockwise, seeing if the gauge is between 57 and 63 cmH2O.
    • Release the control lever and note that the lever moves to the center and pressure drops immediately to 0 cmH2O.
    • If any of these pressures are wrong, see “limited pressure” on the troubleshooting guide.
    • With the control lever on inhale, adjust the pressure knob to 40 cmH2O.
    • Turn the inhale pressure knob fully counterclockwise. Check that the gauge is between 18 and 22 cmH2O.
    • Slowly returning the inhale pressure knob clockwise should increase the pressure gradually back to 40 cmH2O.
    • 5. Turn the pressure knob fully clockwise, set the control lever to inhale, remember the gauge reading, and release the control lever.
    • Set inhale flow to minimum (clockwise). Set the control lever to inhale and check that the pressure is within 10 cmH2O below the previous reading (maximum flow).
    • Remove the rubber plug. Check the flow by pushing the control lever to inhale and feel the airflow out of the unit. Release the control lever.
    • Set the inhale flow knob fully counterclockwise, and repeat the test, checking for a large increase of airflow.
    • (The rest of this guide only applies to the CA series, not the CM series.)
    • Set the pressure knob to minimum. Move the switch from manual to automatic.
    • Verify that the unit immediately begins with the inhale pressure, followed by exhale (negative) pressure then drops to 0 cmH2O and continues to repeat.
    • Set the switch back to manual and verify that the pressure drops to 0 cmH2O. Return to auto.
    • One at a time, set each timing knob to 5 seconds and verify that each knob affects the correct phase.
    • Set the timing knobs to 5 seconds and use a stopwatch to check that each one lasts between 4.8 and 5.2 seconds.
    • Set the knobs to 1 second and verify that each phase lasts between 0.9 and 1.1 seconds.
    • If the timings are off, see “incorrect timings” on the troubleshooting guide.
crwdns2935103:0crwdne2935103:0
Jaycen Kyle Mouillesseaux
crwdns2935283:004/10/20crwdne2935283:0
494 crwdns2915208:0crwdne2915208:0
Cal Poly, Team S11-G18, Paton Spring 2020 crwdns2935289:0Cal Poly, Team S11-G18, Paton Spring 2020crwdne2935289:0
CPSU-PATON-S20S11G18
crwdns2936753:030crwdne2936753:0 8
crwdns2942667:0crwdne2942667:0 407
